measurements to 10 millionths of an inch

alt.machines.cnc, rec.crafts.metalworking measurements to 10 millionths of an inch

Anybody know of a calibration lab or testing facility that can make xyz measurements of a small sample with a max dimension of about 2 inches. Accurracy should be in the +-10 millionths of an inch range.

Hul

Reply to
Hul Tytus
Loading thread data ...

Are you sure about that 10 millionths accuracy spec? The very best CMMs can only do about 14 millionths over that range. If that is sufficient, you need to find someone who has a Zeiss UPMC Ultra CMM. Remember when measuring to this level, you have to consider very carefully the pattern of points to be collected and how the final dimension is calculated. If you are talking about a flat or nearly flat sample with some sort of pattern, there are optical profilometers that may do the job.

Reply to
anorton

Optical machines may be fancier than required in this case. The device to be measured is a 3/8 square bar about 2.5 inches long with some steel balls soldered to one side. The points to be measured are the balls themselves and their size is known. A lab having a flat surface with a post imbedded to hold an indicator capable of comparing at 10 millionths plus a full set of gage blocks would be able to make such measurements. A 1x2x3 type of gage with vice like clamping means would be needed but that should be all - hopefully. I've found one outfit that states 70 millionths, but that's pretty far outside the needs here. The 14 millionths you mention would certainly be close enough and I'll add those devices to the search. Thanks--

Hul

In rec.crafts.metalworking anorton wrote:

Reply to
dbr

wow- how would you adjust the position of one of those balls if it was soldered out of tolerance ? I suppose a misapplied tap could perpetuate to move the others.

Reply to
raamman

te:

You are facing a very tough problem.

That come out to 350 nm. I have done a bit on MEMS work and an optical microscope is going to have problems imaging that small of detail. At that tollaerance one is going to have to specify temprature. For example if the rod is steel one deg F of temprature is going to change the lenth ot the bar 12 mico-inches.

Reply to
toolbreaker

It may be possible to get measurements accurate enough with the simple set-up you describe but do not under estimate the difficulty. To measure the apex of a ball you need a flat anvil on the indicator. But to get 10 millionths accuracy when the anvil contacts the gage block, it would have to be in contact to better than 1 fringe optically. In other words you would really have to wring the anvil onto the gage bock as well as the gage block onto the bottom flat (which would have to be a glass or metal optical flat, not a granite table). This would be tricky. The indicator anvil would also have to be set up perfectly parallel to the bottom flat. Having a very small flat anvil would help a lot. I know I would not trust the measurment of one person or lab for this kind of thing

You might do better using a supermicrometer:

formatting link
while these have a repeatability of 10 microinches, the accuracy is only

  1. There is also a method for measuring balls where you can directly compare the ball to a gage block stack using two optical flats. It might work here if you have a jig to hold the part vertical. The ball is placed between a pair of optical flats on one side, and a gage block on the other side. Measure the spacing of the fringes formed between the block and the upper optical flat and calculate the angle of the upper flat. Then measure the distance between the edge of the block and the apex of the ball to calculate the height difference between the block and the ball.

Reply to
anorton

I ran a mitsubishi wire edm that had a 5th place button. I hit that thing ten times and it moved a total of .0001 of an inch. Go find a shop with one of those.

Reply to
vinny

Those Mitsubishis are crude slagheaps compared to the best Japanese machines- 10 nanometer resolution! (that's less than half a micro-inch).

Anyway, there's a lab in Switzerland that's supposed to be about the best around.

Reply to
Spehro Pefhany

On a related note, to show how far one can go with off-the-shelf gages, you might be interested in a 0.1 micrometer micrometer. :

formatting link
That's about 4 microinches. Not bad. Just don't breathe when you use it. d8-)

Reply to
Ed Huntress

Ed - A local calibration lab has a Mitutoyo tool about the size of a 6 in by 10 lathe with a telescoping "spindle" that touches the tail stock. The travel on the "spindle" is readout in millionth's of an inch. A standard gage block is placed between the spindle & tail stock and nulled then one to be "calibrated" is put in its place. Looks like speed of operation is as much a seller as the resolution. The mic you mentioned comes close, not considering the tiny range and meager 5 millionths resolution... For the job I'm looking at, 10 millionth's resolution is all that's required, ie the objective is to compare with a stack of gage blocks. A tenth's indicating dial indicator comes close, assuming 1/2 a tenth can be eyeballed - that's 50 millionths.

Hul

Ed Huntress wrote:

Reply to
dbr

They have some great stuff. Mitutoyo was my client for years, until they moved their HQ out to Illinois from New Jersey.

Their metrology lab in Japan is something. Three stories underground, with the room suspended on some kind of shock absorbers. When I used to visit NIST's labs in MD, the guy in charge told me that Mitutoyo's lab probably was the best in the world.

Reply to
Ed Huntress

PolyTech Forum website is not affiliated with any of the manufacturers or service providers discussed here. All logos and trade names are the property of their respective owners.