In what way? The ASCII graph had no values associated with it so the actual nonlinearity % is unknown. Generally a load cell plot shows an error from a straight line. Y-Axis being the error as a % of fullscale and X-Axis being the force. If the graph shown was simply load vs output then I agree the error would appear to be rather large.
The original OP's question appears to have never been answered. Normally when a load cell is calibrated the results are normalized with the initial zero reading subtracted such that zero load = zero output. A load cells zero reading is arbitrary since mounting methods etc. can shift the initial zero considerably. A load cell calibration therefore typically shows span.
I work for Interface Inc. who manufactures load cells and have written the calibration software they use and sell.