Beckman HV-211 High Voltage probe?

Does anyone know what are the ratings of a Beckman HV-211 high voltage probe? I found one webpage that says it is for a 22 megohm multimeter, but it did not state its voltage rating.

formatting link
thanks

i
Reply to
Ignoramus25712
Loading thread data ...

For HV gear, if the ratings are not clearly marked on the item I certainly wouldn't trust my safety to it.

Pete C.

Reply to
Pete C.

"Ignoramus25712"

** According to the third post down - it is 50 kV and 1000 megohms.

formatting link
Will give readings about 5% low with a 10 meg DMM.

Easily checked and calibrated.

........ Phil

Reply to
Phil Allison

Thank you Phil. I will make sure to multiply my readings by 1.05. I appreciate your help.

I just tried this probe to measure house mains AC voltage.

AC voltage (real): 251 V Reading using my probe: 0.55V

That would suggest that when measuring with the probe, I need to multiply my readings by about 251/0.55 = 456 times, to get the actual voltage.

Is that the right calculation?

i
Reply to
Ignoramus25712

"Ignoramus25712"

** Something is wrong.

  1. Test the resistance across the twin plug with your ohmmeter - should be about 1 megohm.

  2. Make sure the twin plug in inserted in your DMM the right way around.

  1. Find a source of DC voltage and compare readings with and without the probe.

....... Phil

Reply to
Phil Allison

Yes, it is strange. I tried measuring DC from a military surplus power supply/battery charger.

With multimeter: 127.1 With probe: 0.130

So, with DC, the probe gives a 1:977 reading,

I am not sure how it could be so different with AC.

In any case, I will need to measure DC.

i
Reply to
Ignoramus25712

"Ignoramus25712" <

** So the probe is OK.

** Think about stray capacitance.

Note how a 10:1 scope probe has to be adjusted to match the scope input to avoid HF response errors.

Maybe adding the right value cap across the twin plug pins will fix your probe.

Try 22 pF for a start.

High voltage mica or ceramic of course !

........ Phil

Reply to
Phil Allison

The design specifications are only as accurate as the manufacturing practices used to build the item. Component values and insulation specifications will drift with aging.

The HV-211 HV probe is for use with a voltmeter having a nominal DC voltage input resistance of 22 Mohms.

Beckman was a quality manufacturer, and their test equipment was considered to be some of the best in terms of dependability. Beckman Industrial instruments, later sold/licensed to Emerson Electric, later operated under Beckman again, later absorbed into Wavetek.

Design specs are generally accurate when the equipment is new, or within a reasonable service lifetime. I generally consider the reliability of an aged, used instrument or accessory to be zero, since the history is completely unknown. Until all internal component values are verified, one wouldn't know if components have been abused, damaged, improperly repaired/altered, or substituted. I don't rely on any used test equipment until I've confirmed the performance with reliable equipment.

The Beckman HV-211 is likely to be about 25 years old.

(from the Beckman specification/instruction sheet) Voltage rating: 0 - 50,000 VDC Input resistance: 1000 Mohms nominal Division ratio: 1,000 to 1 Accuracy: (with 22Mohm meter load) +/- 2% at 25kV changing linearly to

+/- 8% at 50kV and 1kV. Maximum input: 50kVDC or peak AC Operating conditions: 0 deg C to 50 deg C, 0% to 80% RH

The internal components of the particular version of the HV-211 that I have, have marked values of 955 Mohm 1% HV resistor assembly (actually

2 joined, large ceramic-bodied HV resistors) and a 1 Mohm 0.25% resistor across the meter terminal connections.

WB ..............

Ignoramus25712 wrote:

Reply to
Wild Bill

I own this probe, bought new in April 1981, and so have the Operator's Manual, a single sheet.

The max voltage is 50 Kilovolts, AC or DC.

It was intended for checking the acceleration voltage on CRT tubes.

It is a 1000:1 divider, designed to work into a 22 Mohm load.

Accuracy on DC +/- 2% up to 25 KV, dropping to +/- 8% from 25 KV to 50 KV. AC won't hurt it or the connected meter, but no accuracy is claimed for AC use.

Joe Gwinn

Reply to
Joseph Gwinn

Thanks Phil. Measuring AC is of relatively little interest to me as of right now. Thanks for your comments. I feel encouraged.

i
Reply to
Ignoramus25712

Thank you WB, I appreciate you taking time to look it up. Precise accuracy is not essential at this point.

i
Reply to
Ignoramus25712

That's great. I have actually a question about its hookup. Does it kind of assume that one side is grounded potential? That it is "safe" to clip the alligator to one of the ends? And then the other end is "dangerous" and high voltage relative to ground, that should only be touched by the tip of the probe?

And if I make some sort of a pulse discharge system, is it safe to ground one side of it, or would it be able to somehow pass possibly harmful oscillations to ground?

Thanks Joseph, that was supremely helpful. I now feel very good about using this probe.

i
Reply to
Ignoramus25712

I think that you are right, although my Extech meter is not the cheapest (ie it cost probably about $300 new). With measured ratio of

977:1, and 50 kVDC isolation, the probe is perfect for my needs. i
Reply to
Ignoramus25712

Very few meters really have anything close to 10M impedance for the AC ranges. Even a HP 34401 is as low as around 2.5M in the AC ranges. I use a Leader AC millivolt meter with a Fluke proble for AC readings. Vacuum tube voltmeters would be suitable as well.

Reply to
Cydrome Leader

The picture that shows it 'slipping under' - Big goof.

The finger goes in front of the large step flange.

The flanges and bumps along the way are impedance pathways - leakage currents are limited. Consider the probe good for 50KV or there abouts.

a 20,000 ohm/volt meter was used with them - so you have a 1000x probe.

Martin

Martin Eastburn @ home at Lions' Lair with our computer lionslair at consolidated dot net NRA LOH & Endowment Member NRA Second Amendment Task Force Charter Founder IHMSA and NRA Metallic Silhouette maker & member

Pete C. wrote:

Reply to
Martin H. Eastburn

It does not assume anything. It's just two resistors in series, built into a HV probe case. The resistor from probe tip to output pin is 999 Mohms, and the resistor from output pin to ground is 1 Mohm.

Output pin and ground are wired to the banana plug that goes to the digital multimeter.

Ground is also wired to the alligator clip lead.

So, if the probe is at high voltage and the ground clip is connected to circuit ground, the output pin is at 0.001 times the input HV.

If that ground clip comes loose, the whole affair is at HV, but through a 1,000 Mohm resistor.

The probe is just a big resistor. Hook it from HV to ground and don't worry about oscillations. Nor can you hurt ground with "harmful oscillations", or anything else.

(1 microfarad)(10^9 ohms)= 1,000 seconds to get to ~1/3 voltage. This will drain the HV capacitors, but very slowly, and maybe not completely. After the voltage stops dropping, connect a much smaller resistor across the capacitor.

It should work just fine for measuring the charge on a HV capacitor.

Although the probe is (was) calibrated for use with a 22 Mohm meter, anything from 10 Mohms up should work just fine, perhaps with a slight error (~5% at 10 Mohm)) due to slight added loading of the 1 Mohm leg of the voltage divider within the probe.

Welcome. But please make sure that the ground clip is quite secure, as if it comes off, your voltage readings will be way low, which could lead to a surprise. It's useful to have a calibration source handy: A

100-volt source will read as 100/1000= 0.1 volts, which is easily measured with a digital multimeter. At 22 KV, trust nothing.

Joseph Gwinn

Reply to
Joseph Gwinn

This information is from the manual that came with my Beckman Model "RMS 3060" multimeter.

HV-211 High-Voltage Probe Specifications:

VOLTAGE RANGE 0 To +/- 50kVdc INPUT RESISTANCE 1000 meg ohms DIVISION RATIO 1000:1 ACCURACY +/- 2% at 25kV, changes linearly to 8% at 50kV MAXIMUM INPUT +/- 50KvDC or peak ac

Reply to
Brooks Kachner

Thanks, that's great... I now know this probe pretty well, it seems.

i

Reply to
Ignoramus20015

What I mean is this: suppose that I have a HV system, such as a charged capacitor bank, such that one side is grounded and another is charged at high voltage.

Suppose also that my body is grounded.

I can measure voltage in two ways:

1) Clip the alligator clip to ground and touch HV side with the pointy end of the probe 2) Clip the alligator clip to HV side and touch the ground with HV probe.

Clearly, 2) is dangerous since my hand, holding the alligator probe, would be almost touching the HV side.

Whereas 1) is not dangerous, as far as I can tell.

If so, then, this probe can be used on HV systems where one side is grounded, in only one way.

This all may sound obvious, but I want top be very redundant in making sure that I use it properly.

i

Reply to
Ignoramus20015

You should run with the cases at a low potential (ground), connect the "hot" end of the probe to the HC terminal on the cap, the alligator clip to the case of the caps, and the other ends to your meter, with the meter set to the right voltage scale.

Most important: Once you have made all of these connections, do NOT touch anything (including the probe OR meter) while there's high voltage on the caps. Do NOT manually "probe" the caps while they are charged. A failure in the high voltage resistor can lead to some nasty problems - and you don't want to be near (or part of) the discharge path if this should occur... =

Reply to
Bert Hickman