UL 94 Flammability Tests

Jun 25, 2004 2 Replies

Hello all!



I'm having some trouble interpreting the meaning of some UL 94 ratings.



The data sheet for our preferred product, which is Makrolon 2658, shows the following ratings for the UL vertical and horizontal tests:



Thickness 0.81 mm UL 94 Class V-2 Thickness 1.5 mm UL 94 Class V-2 Thickness 3.0 mm UL 94 Class HB Thickness 6.0 mm UL 94 Class HB



Although the four lines of data in the table are arranged in ascending order of thickness, they refer to the test conditions that were used on two *different* procedures.



The more stringent test is the vertical one, and a V-2 rating indicates less flammability than V-0. Performance of the material on this test may vary by thickness of the sample, since burn rate and combustibility will be higher for thinner cross-sections. So thicker is better. And other products (eg Makrolon 2667) show a rating that rises with increased thickness.



My challenge is to justify that a 2mm thick instrument housing, fabricated from Makrolon 2658, meets V-2 standards. However, if thicker is *always* better, why would the vendor even publish the numbers for the HB test on this particular polymer? Horizontal tests are typically done with thicker samples than vertical ones anyway, and it would seem to me that the vertical test takes precedence in this case.



Any thoughts on this would be much appreciated.



Thanks!


--------------------------------------------------------------------- John Murray PhD CEng Pemstar Pacific Consultants



2020 South 10th Street San Jose, California 95112
+1-408-282-4805 (office)
+1-408-918-5080 (fax) snipped-for-privacy@pacificconsultants.com

---------------------------------------------------------------------


I think your logic is correct in predicting that the material would meet V-2 at

2mm. However, if engineering certainty is required for a specification, you may have to insist that the actual test be run.

V-0 is actually the best rating in the vertical test. V-2 allows flamming drops.

RT

I believe those numbers mean, that at 3 & 6 mm only HB-tests and at 0,81 & 1,5 mm only V2-tests were done, for certification. If you did the V2-test at 3 & 6 mm you would also pass the V2-rating (in addition to HB). Typically, UL-ratings get worse with decreasing thickness.

"John Murray" schrieb im Newsbeitrag news: snipped-for-privacy@posting.google.com...

Join the Discussion

Have something to add? Share your thoughts — no account required.

Didn't find your answer?